  #16  
Old 07-20-2012, 06:53 AM
Jim_Dont_Care's Avatar
Premium Member

Jim_Dont_Care's Tank Specs
 
Join Date: Mar 2005
Location: grand blanc michigan
Posts: 2,195
Zip Code: 48439
Real Name: jim
My Mood: Cynical
Default

i use fluid but also if i check it with ro\di it comes up zero is all i was saying so its acurate for me
Members of Michigan Reefers have rated post as the most helpful. Skip right to it!
Similar Threads
Thread Thread Starter Forum Replies Last Post
Friendly Reminder: Calibrate Your Refractometer Aegis Reef Discussion 19 01-11-2010 04:42 PM
how to calibrate a pin point ph monitor BOODARE03 Lighting, Filtration and Other Equipment 2 11-13-2009 08:32 PM
How do you calibrate a tds meter? trick440 Lighting, Filtration and Other Equipment 3 08-11-2008 11:25 PM
Do you need to calibrate a new PH Probe? trick440 Lighting, Filtration and Other Equipment 1 07-31-2008 07:16 PM
How often do you calibrate your refractometer? Taylor_made Lighting, Filtration and Other Equipment 2 04-27-2006 07:42 PM

  #17  
Old 07-20-2012, 02:38 PM
woods4542000's Avatar
Member
woods4542000's Tank Specs
 
Join Date: Jun 2009
Location: webberville
Posts: 219
Zip Code: 48892
Real Name: Mike
Default

Quote:
Originally Posted by TG_Lansing View Post
OK, I am also paranoid about my refractometer calibration. I normally use RO/DI as my calibration solution came with ran out. Where can I get a calibration fluid? Any cheap but good places?
You can get it at preuss for like 7$ I think.
  #18  
Old 07-20-2012, 08:39 PM
mf2055's Avatar
Member
mf2055's Tank Specs
 
Join Date: Jan 2004
Location: Grand Rapids (Standale)
Posts: 2,159
Zip Code: 49534
Real Name: Andrew
My Mood: Fine
Default

Quote:
Originally Posted by katymunoz View Post
Ok so now I am truly confused with my calibration!

So I have two solutions and one solution says its 35 ppt and 1.033 refractive index, and the other one says it is 35ppt and 1.026 refractive index...... So how is this possible?? If I calibrate one to 35ppt then the other is off when I put it on my refractometer.... So I officially don't know what my salinity is!
Salinity is temperature dependant.
  #19  
Old 07-21-2012, 09:11 AM
katymunoz's Avatar
Premium Member

katymunoz's Tank Specs
 
Join Date: May 2011
Location: warren
Posts: 1,747
Zip Code: 48091
Real Name: Kathryn
My Mood: Amused
Default

Quote:
Originally Posted by mf2055 View Post
Salinity is temperature dependant.
Both say at 77 degrees right on the bottle. The one says:

53 ms solution. 35ppt, 1.033 refractive index at 77 degrees.


The other bottle says 35ppt, refractive index 1.026 at 77 degrees.

So that's where my question still is, how can two different solutions at 35ppt have different refractive indexes according to the manufacturer.
  #20  
Old 07-21-2012, 11:46 AM
dputt88's Avatar
Member
dputt88's Tank Specs
 
Join Date: Jul 2011
Location: romeo mi
Posts: 1,688
Zip Code: 48065
Real Name: doug
My Mood: Yeehaw
Default

Quote:
Originally Posted by mf2055 View Post
Salinity is temperature dependant.
salinity is not temperature dependent, specific gravity is.

assuming your solution is completely dissolved and it reads 35 parts per thousand, no matter how the temp swings (as long as it doesn't boil) you will always have 35 parts per thousand, unless you found a way to create matter.

specific gravity is temperature dependent because it refers to the density of a solution, not the concentration. as we know at different temp a solution will have a different density (ice floats in water).
  #21  
Old 07-21-2012, 11:50 AM
dputt88's Avatar
Member
dputt88's Tank Specs
 
Join Date: Jul 2011
Location: romeo mi
Posts: 1,688
Zip Code: 48065
Real Name: doug
My Mood: Yeehaw
Default

Quote:
Originally Posted by katymunoz View Post
Both say at 77 degrees right on the bottle. The one says:

53 ms solution. 35ppt, 1.033 refractive index at 77 degrees.


The other bottle says 35ppt, refractive index 1.026 at 77 degrees.

So that's where my question still is, how can two different solutions at 35ppt have different refractive indexes according to the manufacturer.
IMO, the way you can have 2 salinity matched solutions with different SG at the same temp is because salinity only refers to the concentration of salt in the solution, all the other ingredients to the solution may vary between brands, therefore so does the solution density.
  #22  
Old 07-21-2012, 01:31 PM
dputt88's Avatar
Member
dputt88's Tank Specs
 
Join Date: Jul 2011
Location: romeo mi
Posts: 1,688
Zip Code: 48065
Real Name: doug
My Mood: Yeehaw
Default

wow, just realized you were talk refractive index and i was still in SG mode. anyways here a article on refractometers maybe your answer is in there. i tried reading all of it but unfortunately i have to leave the computer now

Refractometers and Salinity Measurement by Randy Holmes-Farley - Reefkeeping.com

also where are you getting those numbers? why are their refractive indices so low? are you sure they aren't 1.33 and the other 1.27.

if so i would use the calibration fluid with 1.33 refractive index because that's the same RI that you'd expect from your tank water.

this is why i use a hydrometer, its a lot easier to understand. lol.
  #23  
Old 07-21-2012, 03:16 PM
katymunoz's Avatar
Premium Member

katymunoz's Tank Specs
 
Join Date: May 2011
Location: warren
Posts: 1,747
Zip Code: 48091
Real Name: Kathryn
My Mood: Amused
Default

Lol yes 1.33 and 1.26 sorry
  #24  
Old 07-24-2012, 12:19 AM
doogle's Avatar
Member
doogle's Tank Specs
 
Join Date: Apr 2012
Location: St. Joseph/Benton Harbor
Posts: 168
Zip Code: 49022
Real Name: Mike
My Mood: Worried
Default

Temp affected specific gravity is a good reason not to use hydrometers. Especially the old ones that are not "temp adjusted". But still.
Reply

Tags
200, apex, cheap, colonies, corals, disease, fish, freshwater, glass, grow, healthy, hydrometer, lansing, led, par, problem, ro/di, salt, sps, tank, tanks, temperature, water, water changes


Currently Active Users Viewing This Thread: 1 (0 members and 1 guests)
 
Thread Tools
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On



All times are GMT -4. The time now is 10:10 PM.


Powered by vBulletin
Copyright ©2000 - 2013, Jelsoft Enterprises Ltd.
SEO by vBSEO
Vendor Tools vBulletin Plugins by Drive Thru Online, Inc.
.
Template-Modifications by TMS
Copyright © MichiganReefers 2003 -